With regard to Ben Wade's MAN plots for F Ka on TAP and the TiO2 outlier, here are some more of my wandering thoughts.
1. The MACs for F Ka in TiO2 are quite large. For F Ka in Ti the tables MACs (from CalcZAF) are:
MAC value for F Ka in Ti = 14588.35 (LINEMU Henke (LBL, 1985) < 10KeV / CITZMU > 10KeV)
MAC value for F Ka in Ti = 14540.00 (CITZMU Heinrich (1966) and Henke and Ebisu (1974))
MAC value for F Ka in Ti = .00 (MCMASTER McMaster (LLL, 1969) (modified by Rivers))
MAC value for F Ka in Ti = 14986.59 (MAC30 Heinrich (Fit to Goldstein tables, 1987))
MAC value for F Ka in Ti = .00 (MACJTA Armstrong (FRAME equations, 1992))
MAC value for F Ka in Ti = 14277.71 (FFAST Chantler (NIST v 2.1, 2005))
MAC value for F Ka in Ti = 14500.00 (USERMAC User Defined MAC Table)
and for F ka in O they are:
MAC value for F Ka in O = 12439.63 (LINEMU Henke (LBL, 1985) < 10KeV / CITZMU > 10KeV)
MAC value for F Ka in O = 12390.00 (CITZMU Heinrich (1966) and Henke and Ebisu (1974))
MAC value for F Ka in O = .00 (MCMASTER McMaster (LLL, 1969) (modified by Rivers))
MAC value for F Ka in O = 11927.75 (MAC30 Heinrich (Fit to Goldstein tables, 1987))
MAC value for F Ka in O = .00 (MACJTA Armstrong (FRAME equations, 1992))
MAC value for F Ka in O = 11863.62 (FFAST Chantler (NIST v 2.1, 2005))
MAC value for F Ka in O = 12400.00 (USERMAC User Defined MAC Table)
I don't have Ben's MDB file handy at the university, but yesterday from home I tried selecting the lowest MAC table values (FFAST), and the MAN fit for TiO2 (#559) was still about the same amount over the trend line. So it doesn't seem possible that the MAC could be off that much.
Interestingly if we look at the calculated MAN absorption corrections as seen here (from running PFE in DebugMode):
MAN fit data for Mg ka, SP4 TAP, 15 keV
Std J MANSet Npts Z-bar Cps AbsCor
504 1 1 1 14.3913 .715813 1.58141
501 2 1 2 10.7106 .523931 1.49128
541 3 1 3 12.0184 .548867 1.43226
559 4 1 4 16.3937 .584618 1.83249 <- TiO2
578 5 1 5 20.5879 .668368 2.41281
701 6 1 6 10.7190 .618357 1.33443
503 7 1 7 13.4430 .615176 1.54753
549 8 1 8 6.12880 .349537 1.25008
571 9 1 9 11.7470 .557961 1.40367
575 10 1 10 10.6460 .638117 1.31164
589 11 1 11 14.0000 .831370 1.09576
594 12 1 12 4.00000 .280255 .964030
602 13 1 13 27.0000 .713668 3.01841
619 14 1 14 21.0000 .834364 1.82118
514 15 1 15 26.6522 .592982 3.27371
we can see that the largest corrections are for F Ka in some of the other standards. So it could be the MAN continuum absorption correction, but why is TiO2 more of an outlier than the other standards with larger corrections?
Note: many years ago I attempted to apply a true continuum absorption correction using Myklebust's continuum expression, but I got worse regression results than just using the normal phi-rho-z bulk absorption correction...