Below are the tungstates and molybdates that I’ve synthesized; they’re listed in chronological order. I used the same alumina crucible for all except the last, for which I used a new alumina crucible. I performed full analyses of the synthetic scheelite and powellite, as I have some excellent, essentially end-member natural material from Bourlamaque, Québec and from the Tonopah Divide Mine, Nevada, respectively. For the other materials, I analyzed for Ca only. Note that measurement of Ca Kα in tungstates is difficult when Ca concentration is low, as W Ll(2) interferes at/near the Ca Kα peak position both on PET and LiF; I used LiFH (sin[θ] = 0.834) and operated in differential mode with a relatively narrow window (2.7 V).
The scheelite appears to be slightly deficient in WO
3 compared to the natural material, however the discrepancy is small and could simply be due to a systematic measurement error. The SrO and MoO
3 appear to be present in the powdered CaWO
4 starting material from Fisher; it's labeled "precipitated pure." I'll plan to grow coarse crystals again and will use CaCO
3 and WO
3 as starting materials instead.
Overall I’m a little disappointed with the results, as I apparently was unable to remove impurities sufficiently from the crucible after each run, even when scrubbing with emery paper. The worst case was the synthetic wulfenite (PbMoO
4), which contains substantial WO
3 -- I didn’t bother to do quantitative analyses of it. I prepared it in a two-step process in order to combat problems with volatility of both PbO and MoO
3: in the first step, I synthesized fine-grained material via solid-state reaction by heating to 1000°C and holding at that temperature for 48 hours. When heated quickly to relatively high temperature, the oxides react quickly enough that little evaporation or sublimation occurs. I then attempted to obtain coarser crystals by dissolving it in Na
2MoO
4 and then cooling slowly. While I was able to dissolve the PbMoO
4 without any problem, it never re-precipitated. However, the grains produced by the solid-state reaction measure up to a few hundred microns across, and so I was able to mount and examine this material without any difficulty.
The ever-present contaminant in the runs is CaO. It appears in concentrations of ~400 ppm in the CdWO
4, which I prepared in a new crucible. I even saw it in one of my CsTiOAsO
4 runs despite starting with a thoroughly cleaned Pt crucible covered with Pt foil. Presumably it is present in the decades-old furnace and begins to mobilize as the temperature is raised.
Another problem I encountered was that the Li
2B
4O
7 flux that I used in the synthesis of Bi
2WO
6 (russellite) reacted readily with the alumina crucible and made it impossible to remove the crystals without breaking the crucible (hence the need for a new crucible for the CdWO
4 synthesis). As you can see below, the Bi
2WO
6 produced some pretty cool textures. The crystals are not acicular, but rather are platy and oriented at high angle to the plane of section.
Unfortunately, my Omega digital temperature controller decided to start spitting out an error code that doesn’t seem to make any sense, especially since the “error” occurs only when the temperature is stable. This forced me to press into service an analog controller that I’ve been building and testing over the past few months. I took the photo below while testing it at 60°C (which gives a reading of ~300 mV) using a hollow, wirewound power resistor as the “furnace,” with the thermocouple probe placed inside the tube. For this relatively responsive “furnace,” I can get the temperature stable to within about 0.2°C with appropriate settings of the error amplifiers. I actually had intended to place this controller in a box, but, as you can see, the control panel managed to outgrow itself.
With the much less responsive real furnace, temperature control is a bit of a challenge, and I’m now rebuilding the circuit to address some difficulties that I came across. The temperature ramp is tricky to construct for good performance, as it relies on exceedingly slow charging of a large polypropylene film capacitor as part of an op amp integrator circuit. The op amp that I’m using is the LMC6081, which has an exceedingly low “typical” input bias current of 10 fA. This means that the op amp inputs can’t come in contact with FR-4 circuit board material, as its surface resistance is too low(!), and wiring needs to be done “dead bug” style in part, with air as the insulator. Further, I’m using universal, breadboard-style prototyping boards, and so component and wiring layout is not optimal overall. The final major problem, possibly related to the former, is that the circuit is oscillating. While the oscillations don’t prevent the circuit from functioning, they are certainly affecting the signal sent to the solid state relay and likely sapping power output. This sort of problem can be really horrendous to deal with, but I’ll have to dive into it soon.