That way one can utilize TDI corrections on more than 5 elements in a sample.
But should you do that?
TDI is correcting for the change in intensity over time.
For the first set of elements, measured over time period 1, TDI is correcting for changes in the X-ray emission of the sample and ideally the regression provides data that correspond to "time zero" - the fresh, unirradiated original composition.
If TDI is applied to a second set of elements measured at the same spot, starting at time period 2, it will only be able to correct back to the composition of the sample as it was at the end of time period 1 = the start of time period 2.
The second application of TDI on the same analytical point is correcting back to an already-damaged/changed/altered material,
not to the original composition.
It may be argued that some further correction is better than none at all, but it must be realized that a second set of TDI corrections during an analytical routine is not a panacea to obtain the actual original composition.
If the sample is sufficiently beam-sensitive that it continues to change during the measurement of a second set of elements,
perhaps a better analytical scheme should be used.
In the case of most zeolite minerals, for which the Fe content is effectively negligible, I find that simultaneous measurement of the 5 main measurable components: Na, Al, Si (all on TAP), K and Ca (both on PET) with a single application of TDI is sufficient.
Cheers, Andrew