While comparing the model to real observed values on the machine, I had found some strange behavior in peaksight (but as PfS is using some dll, from cameca for communication and data transfer from machine it can be affected too) depending from set software dead time.
The dead time and count capping is nicely illustrated with comparison of WDS scans of same substance at different probe currents. It is clear that background is exactly same (As it is low count rate) when comparing with cps/nA, but high current peaks are capped.
From the beginning of SXFiveFE installation in 2014 I was always in impression that our spectrometer 5 has worse performance as amplitude in WDS scans always were lower. With low count rates the difference is not big, but with high current it is like half. While doing 660nA on Cr2O3 I spotted that such discrepancy is obviously not right as xray rate meter showed nearly identical intensity bars while at peaks.
When setting interpretor for do count for 1 second, the counts would result in the very close values. So Quanti is also not affected, but WDS scans are. I tried to acquire mapping of Cr Ka at 660, and the intensity for 5 spectrometer is reduced. As I did few WDS scan runs at different set dead times, I saw that there is no discrepancy when dead time is set to 1µs, The WDS amplitudes of peaks are near identical and intensity mapping is also near identical for both 2 and 5 spectrometers.
This is really enigmatic behavior, had you seen this on PfS? The results probably is differently acquired from machine when doing image and WDS scan (needs to be pushed to software constantly with WDS position/stage position movement), than when using hardware counters for measuring counts in few seconds for quantification.
I see that mappings should be done at 1 µs DT.