We have such spectrometer from 2014 on our SXFiveFE.
I can tell you that I am having love-hate relation with it.
cons:
1) I think the mapping approach (on PC0) is better than Oxygen peak shift, it is not clear how much it is shifted, in case of mixed oxidation states the peak will be broadened. Generally I see O Ka shift to be really complicated to have any use.
2) on extended spectrometer the oxygen Ka is situated at the very edge of "spectrum", it is enough to get some mechanical de-calibration and oxygen peak is no more accessible (until Spectrometer is serviced).
3) Best spectral resolution (?) – only for Oxygen! Please mind, that normally TAP's suffer from too broad peaks - they (mounted diff. crystals) are normally smaller than PET or LiF and is further narrowed by blanking at the edges, so that there would be even narrower peaks and less overlaps. And here you have LTAP - at anything else than Oxygen (and Florine) with complicated composition that crystal is a complete nightmare! Well that can be mitigated a bit if You use ProbeSoftware and can use MAN background (unfortunately, we don't have ProbeSoftware), because it is impossible to find any place for classical background measurements for Si, Al, Mg, M lines of REE, Hf- W, Y, Sr, Rb. That spectrometer is idle for most of our analysis, unless very simple minerals are analysed.
4) If You are going to have LTAP, you can have only one more crystal mounted on the turret, large crystals are mounted only on the double and not quadruple turrets. Those actually are much better designed than new type (2014) quadruple, and so I would not fear for wires. We have this extended spectrometer with LTAP and LPC0. We are often switching and had no wire failure (differently than with quadruple turrets...)
pros:
1) I prefer this crystal over LPC0 for F measurements. Maybe I would have different opinion if I would use Probesoftware and its MAN background modeling, however with classical background measurement LTAP have enough of space (for F) out from interference for background measurement in complicated minerals (containing REE, Fe (try that on LPC0, PHA wont help you there)).
2) I think LTAP can have some potential for measuring Fe L lines for Fe oxidation state determination (it also suffers from similar uncertainties as Oxygen peak shift), That needs to be clarified. As for oxygen I find LPC0 quite impressive for that.
Concluding:
I would prefer instead of current extended spectrometer with x2 turret (LPC0, LTAP) - the spectrometer with x4 turret (TAP, PC0, PET). It would then could work, instead of being 90% of time idle.