I've been running some tests of my own on this question from Rom on his efforts to utilize Fe metal as an Fe primary standard for various Fe oxides where the Fe is present in major concentrations.
As some of you may remember, Rom found that when using Fe metal as a primary standard, he was getting consistently low totals for his Fe oxide standards, but his sulfide standards seem to analyze close to their expected values. I can now report that I am seeing similar effects.
Why could this be? Well as discussed in the posts above, it could be a question of standard accuracy, e.g., his hematite standards could be contaminated with OH or H2O. But I am seeing similar issues with my own magnetite standard and that has been analyzed for Fe using wet chemistry and for ferric-ferrous ratios using colorimetry by Ian Carmichael many years ago.
It is also unlikely to be a matrix correction or background issue, as discussed above. Though it might be worth checking the dead time calibration, though at 20 nA that is unlikely to be a significant effect. So what else could it be?
Now we all know that when analyzing lower Z elements with low energy emission lines where the valence shell is one of the shells involved in the electron transition which produces an x-ray emission, there can be significant peak shape/shift effects from chemical bonding, such that we must usually utilize a primary standard that is at least somewhat similar to our unknown.
For analysis of light elements such as O, N, C or B it is well known that such chemical peak shift/shape effects can be quite large, even when utilizing relatively low resolution modern LDE diffractors. In such cases, the use of a close matrix matched primary standard is necessary, or one can utilize area peak factors (APFs) to account for these chemical bonding effects:
https://probesoftware.com/smf/index.php?topic=536.0One can also utilize integrated intensities for the acquisition of WDS intensities which can be quite slow, but can handle these peak shift/shape effects automatically. See the Integrated Intensities options in the Elements/Cations dialog:
https://probesoftware.com/smf/index.php?topic=536.msg2992#msg2992And even in the case of say Mg, Al and Si Ka, we would not want to utilize a metal primary standard when analyzing those elements in oxides or silicates. Instead we would utilize an oxide or silicate standard such as MgO, MgAl2O4 or Mg2SiO4 for Mg Ka analysis of oxides and silicates. An exact matrix match is not necessary for these emission lines, but we can't reliably extrapolate from a pure metal to the oxide/silicate.
However, for higher Z, higher energy emission lines, e.g., Fe Ka, I would have thought that these chemical peak shift/shape effects would be minimal, but perhaps that is not the case. For example, here is a measurement of Fe Ka using Fe metal as a primary standard, analyzing pyrite as a secondary standard:
St 730 Set 1 Pyrite UC # 21334, Results in Elemental Weight Percents
ELEM: Fe Mn Cr Si S Ti
TYPE: ANAL ANAL SPEC SPEC SPEC SPEC
BGDS: LIN LIN
TIME: 40.00 40.00 --- --- --- ---
BEAM: 29.89 29.89 --- --- --- ---
ELEM: Fe Mn Cr Si S Ti SUM
342 46.309 -.007 .000 .000 53.450 .058 99.810
343 46.229 -.012 .000 .000 53.450 .058 99.725
344 46.329 -.039 .000 .000 53.450 .058 99.797
345 46.304 .004 .000 .000 53.450 .058 99.816
346 46.323 .018 .000 .000 53.450 .058 99.849
AVER: 46.299 -.007 .000 .000 53.450 .058 99.799
SDEV: .040 .021 .000 .000 .000 .000 .046
SERR: .018 .010 .000 .000 .000 .000
%RSD: .09 -288.57 .00 .00 .00 .00
PUBL: 46.550 n.a. .000 n.a. 53.450 .058 100.058
%VAR: -.54 --- .00 --- .00 .00
DIFF: -.251 --- .000 --- .000 .000
STDS: 526 525 --- --- --- ---
Here we can successfully extrapolate from Fe metal to FeS2 (as Rom found). Another attempt:
St 730 Set 2 Pyrite UC # 21334, Results in Elemental Weight Percents
ELEM: Fe Mn Cr Si S Ti
TYPE: ANAL ANAL SPEC SPEC SPEC SPEC
BGDS: LIN LIN
TIME: 40.00 40.00 --- --- --- ---
BEAM: 29.89 29.89 --- --- --- ---
ELEM: Fe Mn Cr Si S Ti SUM
387 46.198 -.032 .000 .000 53.450 .058 99.674
388 46.255 -.034 .000 .000 53.450 .058 99.729
389 46.237 .005 .000 .000 53.450 .058 99.750
390 46.361 .024 .000 .000 53.450 .058 99.893
391 46.218 -.016 .000 .000 53.450 .058 99.710
AVER: 46.254 -.010 .000 .000 53.450 .058 99.751
SDEV: .064 .025 .000 .000 .000 .000 .084
SERR: .029 .011 .000 .000 .000 .000
%RSD: .14 -240.04 .00 .00 .00 .00
PUBL: 46.550 n.a. .000 n.a. 53.450 .058 100.058
%VAR: -.64 --- .00 --- .00 .00
DIFF: -.296 --- .000 --- .000 .000
STDS: 526 525 --- --- --- ---
Again within 1% relative accuracy. Now let's analyze the magnetite standard from Carmichael:
St 395 Set 1 Magnetite U.C. #3380, Results in Elemental Weight Percents
ELEM: Fe Mn Cr Si S Al Mg O
TYPE: ANAL ANAL SPEC SPEC SPEC SPEC SPEC SPEC
BGDS: LIN LIN
TIME: 40.00 40.00 --- --- --- --- --- ---
BEAM: 29.90 29.90 --- --- --- --- --- ---
ELEM: Fe Mn Cr Si S Al Mg O SUM
322 69.969 .046 .007 .000 .000 .201 .072 27.803 98.098
323 69.922 .017 .007 .000 .000 .201 .072 27.803 98.021
324 70.188 .036 .007 .000 .000 .201 .072 27.803 98.307
325 70.542 .056 .007 .000 .000 .201 .072 27.803 98.681
326 70.528 .006 .007 .000 .000 .201 .072 27.803 98.616
AVER: 70.230 .032 .007 .000 .000 .201 .072 27.803 98.345
SDEV: .296 .021 .000 .000 .000 .000 .000 .000 .297
SERR: .133 .009 .000 .000 .000 .000 .000 .000
%RSD: .42 64.40 .00 .00 .00 .00 .00 .00
PUBL: 72.080 .054 .007 .000 n.a. .201 .072 27.803 100.217
%VAR: -2.57 -40.75 .00 .00 --- .00 .00 .00
DIFF: -1.850 -.022 .000 .000 --- .000 .000 .000
STDS: 526 525 --- --- --- --- --- ---
Our relative error has now increased to ~2.5% which is unacceptable. Analyzing other Fe silicate (glass) standards we observe the same low accuracy, here for NIST SRM K-412:
St 160 Set 2 NBS K-412 mineral glass, Results in Elemental Weight Percents
ELEM: Fe Mn Cr Si S Mg Ca Al O
TYPE: ANAL ANAL SPEC SPEC SPEC SPEC SPEC SPEC SPEC
BGDS: LIN LIN
TIME: 40.00 40.00 --- --- --- --- --- --- ---
BEAM: 29.89 29.89 --- --- --- --- --- --- ---
ELEM: Fe Mn Cr Si S Mg Ca Al O SUM
352 7.563 .058 .000 21.199 .000 11.657 10.899 4.906 43.597 99.879
353 7.510 .047 .000 21.199 .000 11.657 10.899 4.906 43.597 99.816
354 7.498 .029 .000 21.199 .000 11.657 10.899 4.906 43.597 99.785
355 7.521 .078 .000 21.199 .000 11.657 10.899 4.906 43.597 99.857
356 7.579 .057 .000 21.199 .000 11.657 10.899 4.906 43.597 99.894
AVER: 7.534 .054 .000 21.199 .000 11.657 10.899 4.906 43.597 99.846
SDEV: .035 .018 .000 .000 .000 .000 .000 .000 .000 .045
SERR: .016 .008 .000 .000 .000 .000 .000 .000 .000
%RSD: .46 33.33 .00 .00 .00 .00 .00 .00 .00
PUBL: 7.742 .077 n.a. 21.199 n.a. 11.657 10.899 4.906 43.597 100.077
%VAR: -2.69 -30.21 --- .00 --- .00 .00 .00 .00
DIFF: -.208 -.023 --- .000 --- .000 .000 .000 .000
STDS: 526 525 --- --- --- --- --- --- ---
Again about 2.5% relative low for Fe. I did not have a Mn sulfide standard, but running Mn metal against MnO, I see a similar low accuracy:
St 25 Set 3 MnO synthetic, Results in Elemental Weight Percents
ELEM: Fe Mn Cr Si S O
TYPE: ANAL ANAL SPEC SPEC SPEC SPEC
BGDS: LIN LIN
TIME: 40.00 40.00 --- --- --- ---
BEAM: 29.89 29.89 --- --- --- ---
ELEM: Fe Mn Cr Si S O SUM
347 .008 74.428 .000 .000 .000 22.554 96.990
348 .015 75.138 .000 .000 .000 22.554 97.707
349 .024 74.848 .000 .000 .000 22.554 97.426
350 .026 74.670 .000 .000 .000 22.554 97.250
351 -.002 74.831 .000 .000 .000 22.554 97.384
AVER: .014 74.783 .000 .000 .000 22.554 97.351
SDEV: .012 .261 .000 .000 .000 .000 .262
SERR: .005 .117 .000 .000 .000 .000
%RSD: 81.45 .35 .00 .00 .00 .00
PUBL: n.a. 77.446 n.a. n.a. n.a. 22.554 100.000
%VAR: --- -3.44 --- --- --- .00
DIFF: --- -2.663 --- --- --- .000
STDS: 526 525 --- --- --- ---
About a 3.5% relative error. And here for Mn2SiO4 synthetic olivine:
St 275 Set 2 Mn2SiO4 (manganese olivine) synthetic, Results in Elemental Weight Percents
ELEM: Fe Mn Cr Si S O
TYPE: ANAL ANAL SPEC SPEC SPEC SPEC
BGDS: LIN LIN
TIME: 40.00 40.00 --- --- --- ---
BEAM: 29.89 29.89 --- --- --- ---
ELEM: Fe Mn Cr Si S O SUM
362 .019 52.526 .000 13.907 .000 31.688 98.140
363 .013 52.363 .000 13.907 .000 31.688 97.971
364 .008 52.843 .000 13.907 .000 31.688 98.446
365 .013 52.404 .000 13.907 .000 31.688 98.013
366 .016 52.316 .000 13.907 .000 31.688 97.927
AVER: .014 52.490 .000 13.907 .000 31.688 98.099
SDEV: .004 .212 .000 .000 .000 .000 .209
SERR: .002 .095 .000 .000 .000 .000
%RSD: 28.58 .40 .00 .00 .00 .00
PUBL: .000 54.406 .000 13.907 .000 31.688 100.001
%VAR: .00 -3.52 .00 .00 .00 .00
DIFF: .000 -1.916 .000 .000 .000 .000
STDS: 526 525 --- --- --- ---
Again about 3.5% relative accuracy. So not good for major element accuracy!
Personally I've always utilized pure oxide standards for my primary standards for measurements in oxides and silicates, and so have never observed these effects previously. Has anyone else looked at these peak shift/shape effects for the first transition series elements? Please share some data with us...
Could these low values be from a subtle chemical peak shape/shift effect? I am running some detailed wavescans to check... but I suspect it's the most likely explanation.
But again, remember that these subtle primary standard matrix effects are something that may be important for major elements, but probably not for trace elements as discussed here:
https://probesoftware.com/smf/index.php?topic=610.msg11752#msg11752A 2 or 3% relative error at 100 PPM is going to produce an absolute error of 2 or 3 PPM, generally not something we are concerned with. Even at 1000 PPM that's only an absolute error of 20 or 30 PPM, so still below the detection limit for many measurements. In other words if you want to measure some first series transition elements at trace levels in an oxide or silicate matrix, and you don't have a suitable pure oxide standard, you are probably OK to use a pure metal standard.