Author Topic: Synthetic Quartz Standard?  (Read 13785 times)
UT E-Beam Lab
    • UT Austin Electron Microbeam Laboratories
« on: December 20, 2013, 08:19:52 AM »
Can anyone recommend a manufacturer where I can purchase a synthetic quartz standard in order to do Donovan's "blank correction" method for measuring Ti concentations in quartz grains?
Many thanks
UofO EPMA Lab
Re: Synthetic Quartz Standard?
« Reply #1 on: December 20, 2013, 09:53:13 AM »
I got my original SiO2 synthetic material at ESPI, now (amazingly enough) located in Ashland, Oregon!
http://www.espimetals.com/index.php/msds/245-silicon-oxide-sio2
They might already have characterized the trace Ti and/or Al. The lot I bought 20 years ago had 1.42 PPM Ti by ICP-MS (A. Konig).
The SiO2 they currently have is probably different material but here is what I have on the original material from 20 years ago:
Specimen from ESPI, 99.99%, EPMA (UCB): Al2O3 ~ 0.01%
Catalog #K4699M
ICP-MS (Alan Konig, PPM) Ti=1.42, Li=3.18, Cu=1.26, Mn=2.93, Pb=0.11, Zn=2.09
Flame AA (Chris Lewis, PPM):
Al=15 +/- 5
Fe=6 +/- 3
Mn=1.5 +/- 0.3
Na=5 +/- 3
Li= 2.3 +/- 0.2
« Last Edit: December 20, 2013, 09:55:11 AM by John Donovan »
UofO MicroAnalytical Facility
UT E-Beam Lab
« Reply #2 on: December 20, 2013, 12:08:29 PM »
Great! Thank you so much
« Reply #3 on: January 13, 2014, 10:41:10 PM »
We use Saint Gobain Spectrosil
« Reply #4 on: January 14, 2014, 02:55:52 PM »
Could you post the trace element characterization and how it was performed on this material?
« Reply #5 on: January 14, 2014, 04:33:37 PM »
I've attached an info sheet with the specs. <10 ppb typical for most elements, but not sure how that was determined. Our LA-ICP-MS people are using it too and it seems to be clean, certainly by EPMA standards.
« Reply #6 on: January 15, 2014, 02:42:45 PM »
This is a very interesting idea to utilize an ultra pure SiO2 glass as a "blank" standard for quartz analyses.
In fact, one advantage of using SiO2 glass is that it is actually *less* sensitive to the electron beam than SiO2 crystal. Here is a reflected light image of a scanned electron beam on SiO2 crystal:
and here is SiO2 glass under the same scan conditions:
But maybe this is not a good thing in that the SiO2 glass might no longer be a similar enough matrix for calibration of the "blank" concentration?
I suppose we'll need to do some high precision measurements and compare traces in SiO2 glass and SiO2 crystal...
In the meantime, I wanted ask everyone: is there a similar ultra pure titania (TiO2) glass that could also be utilized as a blank standard for traces in rutile?
« Reply #7 on: January 15, 2014, 04:12:57 PM »
I don't know - I would expect the beam sensitivity to vary a bit between quartz samples as well?
It does seem to work for us - We've put our Hot Springs Quartz chip under the LA-ICP-MS and got 1.2 ppm Ti. With the probe our long term average is 3+/-3 ppm using the blank correction, which is +16+/-2ppm.
« Reply #8 on: February 05, 2014, 08:22:35 PM »
Late reply...
A lot of XRD uses Spectrosil qtz glass tubes to put their samples in. If you have a friendly XRD tech they might be happy to smash one for you. I obtained a few tubes and analysed fragments from each one for Ti (+Al and REE) via LA-ICP-MS. The resulting average Ti concs. from each tube were well within error of each other (~4ppm). So although it sounds like my glass is a little less pure than some others Spectrosil, it doesn't really matter what it is as long as its well characterised.
At the microprobe workshop Karsten just ran here in Adelaide, I gave Dave Adam's a small chip of synthetic TiO2. This material was given to me on request to our local museum (and apparently was quite expensive for them to obtain), but I would be willing to send you a small chip. However....there was no information on its trace composition, AND I haven't had a chance to analyse it myself. So it could be poor material, I just don't know. If you are willing to take that chance I could send you some.
« Reply #9 on: March 03, 2014, 01:07:04 PM »
Yes, that is correct.
Because the blank correction is fully quantitative it can handle the situation where one's "blank" standard contains a non-zero concentration of the element of interest, as long as that non-zero value is accurate. The reason it's called a "blank correction" is simply because it's usually easier to find a pure material with essentially a zero concentration as opposed to a standard containing a known non-zero concentration.  But both concentrations (zero and non-zero) are equally good at ascertaining accuracy!
Here is a poster and PPT with the blank correction equation attached. Note that this correction must be incorporated into the matrix iteration to achieve the necessary accuracy for trace element characterization:
« Reply #10 on: January 03, 2018, 03:01:33 PM »
Here's a plot I made a while ago comparing quartz glass and crystal for Si Ka and O ka.  Note the absorbed current changes.
Note that over time the oxygen intensity goes up in the crystal, while it goes down in the glass, though more slowly. I have no idea what this means... but thank goodness for the TDI correction.
« Last Edit: April 13, 2020, 10:13:48 PM by John Donovan »
